Calibration Best Practice
Posted: Thu Feb 02, 2017 12:50 pm
I started at a new company and their QC dept. operates calibrations in a different way than I'm used to.
They run STD1, STD1, STD2. They check if the two STD1's are in agreement and if so proceed. If STD2 is in agreement with the second STD1 then proceed. The avearge of the response factor for the second STD1 and STD2 is calculated and this is used in the calibration table.
How do people feel about this??
I would normally have SST standards, Calibration standards, and then check standards. I would only ever calibrate from the calibration standards (there many be multiple of them though) and I feel like, above, STD2 is a check std.
Cheers
Dan
Cheers
Dan