HCG
5 posts Page 1 of 1
HCG
i would like to analyse HCG harmone by hplc for assay, any tips
no really similar, i would like to analyse using UV for assay without denaturing
If it is a drug product try 214 nm where the peptide bond absorbs. If it is a blood/plasma sample you're stuck with LC/MS/MS.
It ALL depends on the expected concentration (ppm versus ppt) and the sample matrix!
Column, mobile phase, and other chromatographic conditions remain the same.
5 posts Page 1 of 1