Reporting "<" values in ChemStation
Discussions about chromatography data systems, LIMS, controllers, computer issues and related topics.
3 posts Page 1 of 1
With an intelligent report, is there a way to add conditional formatting field so that any values below calibration report as "<LOQ" or a similar text?
You need to create an IF statement in the value field
=iif(compound_concentration<value,"<LOQ",compound_concentration)
Of course you can replace Copmpound_concentration by Compound_Amount or Area or whatsoever.
iif is not a typo! it has to be that way
Freek Varossieau
OpenLab CDS 2 specialist
BeyondOpenLab
beyondopenlab@gmail.com
+5977114721
Thank you!
3 posts Page 1 of 1
In total there are 2 users online :: 1 registered, 0 hidden and 1 guest (based on users active over the past 5 minutes)