Page 1 of 1
GC detector configuration broblem
Posted: Wed Jun 28, 2017 12:08 pm
by y.ALKHADIRI
hi everyone,
i am a GC user , recently i confront an error message that says "detector broken" or "signal configuration mismatch". could any one assist me with help and what i am supposed to do ?
Re: GC detector configuration broblem
Posted: Wed Jun 28, 2017 12:21 pm
by tkubowicz
Hello
What you supposed to do first is to give us more details...Different GC systems gives different errors.
Give us details about:
1.GC model information
2.Type of detector
3.Software for instrument control
regards
Tomasz Kubowicz
Re: GC detector configuration broblem
Posted: Wed Jun 28, 2017 12:32 pm
by y.ALKHADIRI
Hello
What you supposed to do first is to give us more details...Different GC systems gives different errors.
Give us details about:
1.GC model information
2.Type of detector
3.Software for instrument control
regards
Tomasz Kubowicz
thank you for your reply,
1.GC model information: 7890B agilent
2.Type of detector : FID
3.Software for instrument control/ OpenLab CDS
Re: GC detector configuration broblem
Posted: Wed Jun 28, 2017 1:00 pm
by tkubowicz
Hello
1. Check FID configuration on GC display (click CONFIG and then DET)
2. Check if you can control temperature, flows, electrometer
3. Try to load DEF_GC.M (default method) in OpenLAb and create new method from scratch - see if you can run method
4.If not working - you can remove FID from GC configuration and reconfigure it.
Regards
Tomasz Kubowicz
Re: GC detector configuration broblem
Posted: Wed Jun 28, 2017 1:37 pm
by y.ALKHADIRI
I Tried to load DEF_GC.M (default method) in OpenLAb and create new method but the detector temperature is stuck on 45 °C ,i configure it to 220°C but it ain't increase.
the only modified parameter from Detector configuration is "column comp"
Re: GC detector configuration broblem
Posted: Wed Jun 28, 2017 1:43 pm
by tkubowicz
Hello
In this case I'd check detector board and GC board (voltage). If you've never done it - It is better call for service.
Regards
Tomasz Kubowicz
Re: GC detector configuration broblem
Posted: Wed Jun 28, 2017 5:13 pm
by uzman
Cartridge heater in the detector , ( or the cables ) seems broken ( open circuit ).
If you have an ohm-meter , you may check the resistance ,