32 karat Software Beckman Coulter HPLC. Failure of detector
3 posts Page 1 of 1
I have a problem with the HPLC (32 karat software). I cannot calibrate the detector it always shows "calibration failed". It was working fine before, with minor problems but it's the first time I see this message.
Thank you for your help
Kaoutar
Do you mean the software was working before but with minor problems or that the detector was working before but with minor problems.
Which detector are you using?
I suspect that the problem is in the detector and not the software. Please post more information.
Also post this in the Liquid Chromatography section.
Kaoutar
Hi
Excuse me, there is any possibility you can share 32 karat software with us?
I am student of PhD in The National Polytechnic Institute in Mexico
I pay any cost of ship
Thanks so much
Best regards
Oscar Guzmán
Excuseme, there is any possibility you can share 32 karat software?