TCD for analysis of H2O, H2S, SO2, etc?
I'm currently running an HP 5890A with capillary column and TCD for analysis of permanent gases (H2, O2, N2, CO, CH4, CO2, etc)
However, I would also be interested in analyzing for: H2O, H2S, SO2, SO3, and possibly even NOx's.
Obviously I'll need a different column, such as the Supel-Q PLOT (not sure if this can elute water?). If I use helium as a carrier gas, would it be possible to detect ppm or low % quantities of these gases with a TCD? Theoretically it looks possible, but I'd like to hear what others use for this analysis, both columns and detectors.
Thanks!
Supel-Q does elute water but it is not the best column to perform water measurement.
The other compounds you mentioned:
'H2S, SO2, SO3, and possibly even NOx's.'
are not measured reliably except with special columns designed to perform that measurement.
N2O is not a problem but NO and NO2 are.
best wishes,
The problem is that the reactive gases ( NO, NO2, H2S, etc ) can react with your detector ( if they don't react with something prior ), and the TCD is usually not sensitive enough, and poisons easily.
Traditionally, chemiluminescent detectors were used for NOx, rather than GC. I suspect that, these days, some of the gases can be more conveniently analysed by other techniques. If you only have to analyse a few samples, and sample size isn't an issue, Drager tubes give quick and dirty results, or submit sample to a environmental lab routinely doing the tests.
Bruce Hamilton