HP 1050 HPLC Autosampler: plunger home not found
Posted: Wed Jul 12, 2017 5:08 pm
Hello-
I tried searching the forum for the answer to this question and wasn't able to come up with anything. I'm getting a little desperate so any help you can provide would be appreciated. I'm going to be a bit verbose, in the hopes that more info can give more insight to my problem.
I have two HP 1050 HPLC systems; they are both very old, but sturdy machines that I believe still have some life in them with the proper care. So I'm fully prepared to take the blame if this is my fault.
We moved office locations recently, so my machine room had to be completely broken down, dried, moved, and reassembled (
). Purging the lines took forever, and once that was all set up, one system works great. The second system does not. These are independent HPLC systems, each run off a separate computer.
For the injectors, I have a dental air compressor with a 2-way split valve to run both of the autosamplers for the two units (it is set to provide 5 bars of pressure). This set up worked fine at the old office.
Now, when the "not good" unit starts up, the normal exhaust sound that happens after the needle moves up and down is so WEAK. Instead of that strong PPPFFFT sound, I'm getting more of a gasp, 'eh,' and the error returned is E22: Plunger home not found. The needle seems to move okay, and I even dug out a spare autosampler and swapped the air valve solenoid from the spare unit - still having the same error. I don't see/hear any air leaks anywhere, I tried swapping the air lines from each machine and the "good" one runs the injector set up just fine with a great exhaust sound.
I'm attempting to go through the manual and run the service tests using the keypad on the unit, but that's not seeming to be too helpful. I'm more inclined to think that this is some sort of electrical/internal programming error, since it wasn't shaken up too much during the move, I doubt it's loose hardware. Just to be safe, I checked all the connections to the board just behind the needle motor and it all seems secure.
Any ideas on what I can do to diagnose the problem? I'm not ready to give up on my ol' buddy just yet!