How do you "activate" a custom field?
I have created a custom field which calculates the percent diff between areas of two injections of a sample and i have the label for this sample in the custom field calculation but how do i activate it to get a result? The custom field is in the project but do i just put Summarize Custom Fields directly under this standard line or does it go at end of sample set?
1) Include output for the CF in a report.
2) The Summarize CFs command usually goes at the end of a SSM.
DR
DR
Hi ,
As DR has mentioned
Make sure you have Calibrate (Label for ref stds) and Quantitate (Label for samples) also at the end of sequence (Processing 'Normal'
As DR has mentioned
Thank you. So if i had 2 summarize custom fields in the one sample set is it sufficient to put "Summarize custom fields" as the last line in the sample set or would i need to add this directly underneath each relevant line? Also do i have to add calibrate and quantitate, i would have thought this was covered in Summarize custom fields?
Hi ,
Last line or at the end of the sample set
If you have created a sample set in the wizard you should have Calibrate and quantitate at the end of the sample set
You need all 3 to produce meaningful results
Calibrate - for standard curve
Quantitate - for unknown (samples)
Typically at the end of the sequence :
Equilibrate , no label , Normal
Calibrate , label for stds, Normal
Quantitate, label for samples, Normal
Summarize Custom Fields, no label , Normal
Hope this helps
Hope this helps
Thanks very much for your reply I will try that.