Fluorescamine precolumn derivatization using Agilent 1100
Can anyone suggest a protocol for automatic precolumn derivatization of primary amines using the Agilent 1100 autosampler?
The one I've been using (below) is not giving good linearity and I dont know if it is due to the program I've developed (guessed!)
John
Draw 0.5ul from buffer (borate pH 10)
Draw 0.5ul from sample
Mix 1.2ul 6 times in air
Draw 0.5ul from fluorescamine (in acetonitrile ~0.4mg/ml)
Mix 2ul 8 times in air
Draw 0ul from wash vial
Wait 2 min
Inject
Hi John
What is the concentration range ?
How good is the repeatability at any given concentration ?
Is the calibration curved, or just scattered ?
Is the deviation from linearity at high or low concentrations ?
My first impression is that the volumes are all small, and therefore vulnerable to large % variability.
Thanks for that Peter.
I've upped the volumes by a factor of 20 and will see how I get on.
It's a first shot in a troubleshooting exercise. I won't ahve time to do a full validation I'm afraid.
Cheers
John