Page 1 of 1

Agilent 1260 infinity DAD lamp without tag question

Posted: Tue Aug 15, 2017 10:41 am
by skooma
Hello!

I would like to install a new lamp which doesn't have a tag on it for Agilent 1260 infinity DAD. From the controller the DAD has following settings:

1)use lamp tag settings(instrument gives an error that the tag is missing and won't start with a new lamp that doesn't have a tag)
2)5190-0917 with cont. heating
3)5190-0917 w/o cont. heating

I'm not sure how to choose between option 2) and 3). What does this cont. heating do and shoud I use it or not.

Re: Agilent 1260 infinity DAD lamp without tag question

Posted: Wed Aug 16, 2017 9:15 am
by leadazide
Try using LabAdvisor to see if it is possible to toggle Lamp Tag required off. On some units this is possible.

Re: Agilent 1260 infinity DAD lamp without tag question

Posted: Wed Aug 16, 2017 5:11 pm
by dap
As far as i remember only one mode will make a lamp ignitable. Another one will cause an error.

Re: Agilent 1260 infinity DAD lamp without tag question

Posted: Thu Aug 17, 2017 7:46 am
by tkubowicz
Hello

You can use lamp without tag. You just need to disable "tag"-recognizing lamp type in DAD config.

Regards

Tomasz Kubowicz

Re: Agilent 1260 infinity DAD lamp without tag question

Posted: Fri Aug 18, 2017 5:21 am
by dap
Image