Empower reports
Posted: Thu Sep 28, 2017 9:53 am
by ydna1977
Hi ,
I have created a number of chromatogram reports within empower for individual injections ie full scale , zoom window , peak results.
I have also created a separate report summarising the results for system suitability and sample summary
My question is how can you combine these into one report ?
Default method properties don't allow this
report type is either summary - individual , summary by all for example
Is it a case of choosing each report and electronic signing for each chromatogram then signing summary report ?
Andy
Re: Empower reports
Posted: Fri Sep 29, 2017 5:00 pm
by DR
open a report, zoom out as far as you can, select everything, copy to clipboard, close report, open an different one, paste at the bottom, rearrange stuff, Save As something else, repeat if needed, rearrange objects to your liking...
Clipboard is your friend in Empower's report editor... Especially when you want the same report contents in both FAT and nonFAT projects :)
Re: Empower reports
Posted: Tue Oct 03, 2017 6:18 am
by ydna1977
Hi ,
Sounds abit long winded to me but I will give it ago
I think I have solved it though
Report must be summary by all , create summary tables at the beginning or end (SST , peak performance , sample summaries ) then add chromatograms
Group chromatograms and peak results and group by date acquired