Mix in hight pressure or low pressure
Hi at All
I have to buy a HPLC with quaternary pump and I don't know what is better because the agilent 1200 is low pressure then Varian hight pressure. What do you think is better?
The Varian gradient pump design is somewhat unique: it has multiple inlet valves in one high-pressure pump head, one valve per solvent. This is not what most people mean by "high-pressure" mixing. Mixing still takes place at low pressure, but there is no separate proportioning valve. Mixing takes place inside the pump head during the fill stroke.
The Agilent uses more the classic design: a quadruple proportioning valve assembly mixes the solvents before they enter the high-pressure pump head. Dionex uses this design too, but I think we did it a little better.
Both designs can deliver excellent results.
Most people use the term "high-pressure mixing" to refer to a different design: a separate pump for each solvent, and the mixer is after the pumps. The pumps change flow rates to make the gradient. The mixing volume for this design can be very small. Agilent (and Dionex) offer this as a more expensive option.
What you want to look at is the gradient delay volume. This takes into account all the various volumes between the point of mixing and the top of the column. (Of course for an isocratic method this does not matter much.) For standard 1 mL/min gradient applications, it usually makes little difference. If you go to 0.2 mL/min then the difference becomes much more important.
Mark Tracy
Senior Chemist
Dionex Corp.
I'm not sure about 1200s but w/ 1100s, you could go 4 lines into 1 pump or 2 lines into 2 pumps.
DR
The Agilent design with 2 lines into 2 pumps is limited to A-B binary gradients, but you can select A1/A2 or B1/B2. What you can't do is form a gradient with one pump, nor can you mix a ternary gradient. I can't think of why this should be impossible, but that is the way it is.
Mark Tracy
Senior Chemist
Dionex Corp.