Cleaning C18 column of peptides
I mistakenly injected high amount of high concentration tryptic digest of protein mixture. Now in every blank I see a huge chromatographic peak. This is a peptide of MW 3403 Da, confirmed from mass spectrometry data. I have washed the column by 95% Acetonitrile for overnight. Even then I see the same peak. The most surprising part is that the peak intensity is increasing! Any solution?
Hello. It seems that mobile phase is better solvent than 95 % AcN. Could the column be washed 1) with the best solvent 2) reversed?
Best regards,
Dmitriy A. Perlow
Dmitriy A. Perlow
Thank you very much. By mobile phase dd you mean water: ACN = 98:2? And what is the best solvent?
It depends on the analyte's nature. I suppose that 95 % AcN is a bad solvent for most of charged polar analytes such as aminoacids or peptides. By mobile phase I mean the mobile phase of the method is used when "the peak intensity is increasing".
Best regards,
Dmitriy A. Perlow
Dmitriy A. Perlow
Hi, Dap. I am running a simple gradient of Water and ACN. It starts with 2% ACN and in 60 minutes it reaches 95% ACN. The peak I'm seeing is around 60% ACN.
Are you suggesting me to wash the column with that particular % of ACN where the peak is eluting?
Yep, it looks a good idea.
Best regards,
Dmitriy A. Perlow
Dmitriy A. Perlow
Thanks a lot Dap. Your suggestion helped me get out of the trouble. What I did:
Disconnected from MS. Increased the flow to 1000 uL/min and started with 5% ACN in water. Passed 10 column volumes and then brought down the flow to 200 uL/min. Connected to MS and checked the spectra. Repeated this step with subsequent increase in ACN, 10, 20, 30, etc. up to 95%. I saw the peptide is coming out at 20% ACN and as I increase the ACN percentage, the intensity of the peptide becomes less. So, I washed the column with 20% ACN in water at 1000 uL/min for 2 hours. When I ran the next blank the peak is now very less. That means this method works. I'm going to wash a bit more and see the blank again.
Thanks again.
Thanks again.
Welcome!
Best regards,
Dmitriy A. Perlow
Dmitriy A. Perlow