Re! Limit of detection for Bromo Benzene by HS-GC analysis
I am just looking for possible " lower Limit of detection for Bromo Benzene by Headspace GC"analysis along with the GC parameters.
Thanks[/b]
I'm sorry to sound pedantic but there are a lot of "it depends".
For example;
Static or dynamic headspace?
What is the sample matrix?
How much sample have you got?
What is your detector?
As an example, and without having done it, I would expect to be able to see 100ppb or lower using SPME/GCMS-SIM on 1g sample heated to 50°C, and sampled for 20mins using a standard addition for quantitation, analysed on a wax column starting at 40°C.
Ralph
The system config as floows.,
Dynamic headspace (G1888 - Agilent)
FID detector (Agilent GC 6890n )
Sample concentration about 100 mg/ml (expected bromo benzene level is about 0.5 ppm )
Sample matrix contains Methanol , Acetone and Bromo Benzene which are used in the process.
We are interested to carry the analysis on GC Capillary columns.
Be aware that it is possible to de-brominate your analyte. Given active sites in your headspace system, reactions can take place and you can easily loose low levels of the bromobenzene.
good luck in your analysis.
Users browsing this forum: Bing [Bot], Semrush [Bot] and 11 guests
