huge hump in the chromatogram
8 posts Page 1 of 1
I am analysing drug in capsule(gelatin capsule) and getting a huge hump before the parent peak. The method works fine for the drug substance. the hump is coming where the two impurities are supposed to be. My mobile phase A is 5/95/0.1:ACN/H2O/TFA, mobile phase B is 95/5/0/1:ACN/H2O/TFA. The Column is C18 and the gradient is
Initial---------5.6%B
10.0 Min.-----50%B
15.0 Min.-----100%B
Please advice how can I get rid of the hump. Thanks in advance.
Have you run a blank gradient? If not, TFA is notorious for causing baseline problems. Read a little more about the reasons in this post:
http://www.sepsci.com/chromforum/viewto ... hlight=tfa
Water check:
Let sit pumping at initial conditions for ~20 min, then run back-to-back blank injections. If the hump is a lot smaller in the second blank, you need to clean up your water, then remake the A phase.
For details, search "Empore"
DR
When I run the drug substance only with same column, mobile phase, gradient etc. I don't see the hump at all. it is when I run the drug in capsule that I start to see the hump and it grows with each injection.
Is there dissolved gelatin in the drug sample? Run a gelatin sample and see if that is the source of your hump. If so, you may need additional cleanup prior to analysis.
Mark Tracy
Senior Chemist
Dionex Corp.
yes! I filtered the sample and also centrifuged it and by doing that I get a clear solution but It is causing that hump at 4-6min. where two of my impurities are eluting. If anyone had similar problem please let me know how to solve this problem. Thanks!
What is in your capsule besides drug? if your standard run is Ok , it indicates that your sample containes some other excipients or your drug interecting with geletin . run one placebo and check.
JM
Dissolved gelatin will pass through an ordinary filter. They are kind of expensive, but centrifugal ultrafilters with a 3000 MW cutoff will remove dissolved proteins. Grace and Millipore are two sources. Another alternative is an SPE cleanup, but that is actually more work than the ultrafilter.
Mark Tracy
Senior Chemist
Dionex Corp.
8 posts Page 1 of 1
In total there are 19 users online :: 2 registered, 0 hidden and 17 guests (based on users active over the past 5 minutes)
Users browsing this forum: Bing [Bot], Google [Bot] and 17 guests