Cresol/Xylene analysis by GC/FID...need help
Hi!
I have to analyse cresol and xylene by GC/FID and splitless injection mode. I use an HP-INNOWax (PEG) column.
Can you send me a good method for the analysis?
What is the best temperature programmation for? And which solvents and internal standard can I use?
Thank you!
Generally beginning at 60°C and programming the oven at 4°C until your peaks elute will work.
Almost any solvent is recommended except water or methanol.
I would not use an injector, column oven, or detector temperature far above 200°C to avoid damaging the column.
good luck and best wishes,
Use a carrier gas oxygen trap with PEG columns.
And which internal standard is good for the method?
The internal standard should chemically be similar to the analytes you are measuring.
In your case diethylbenzene? Cumene? Even Toluene?
You could also use a simple alkane if that is all you have available.
best wishes,
The producer of the column wrote me that I can use any solvent I want. Also water and methanol! So, the best choise would be water as solvent.
What do you think about it?
I would NOT use water or methanol if possible. Ether would be a possibility if you need to make sure any trace water is dissolved.
I would also avoid acetone but might consider 2-butanone.
I WOULD USE
any hydrocarbon, aromatic, or chlorinated hydrocarbon solvent.
For example: pentane, hexane, heptane, tetraethylbenzene, methylene chloride, chloroform, etc
Be certain that cresol and xylene do not coelute with the solvent you pick.
best wishes,
Users browsing this forum: Bing [Bot], DuckDuckGo [Bot] and 9 guests