kf auto titrator calibration
can anybody tell me why we are using sodium tartrate di hydrate during KF auto titrator calibration?
Because it is not hygroscopic and contains a known, stable amount of water.
As Tom said, it's a primary standard ( water content 15.66% mass/mass ), it's readily soluble in methanol, plus it's the suggested calibrant by the USP for samples with trace amounts of water ( < 1% ), although you can also use water.
Bruce Hamilton)
THANKS BOTH OF YOU.
We use water as the KF calibrant. We are usually looking at higher levels, from a few percent to 90%. We use a gas-tight syringe, 100 ul.
In total there are 7 users online :: 3 registered, 0 hidden and 4 guests (based on users active over the past 5 minutes)
Users browsing this forum: Bing [Bot], Google [Bot], Semrush [Bot] and 4 guests