H2S in gas sample
2 posts Page 1 of 1
I have to make injections in porapakQ-TCD with a sample with CO2, CH4, H2, and it maybe contains small quantities of H2S. Is H2S poisonous for porapakQ column or TCD? which temperature must I set the column for H2S to elute?
tolisfoto,
H2S will be OK on a PorapakQ column and a TCD. An old Chrompak application shows on a 9 Ft PorapakQ column at 50'C, and helium carrier gas at 30 ml/min, CO2 has a RT of 1.5 mins and H2S has 5.2 mins.
Gasman
2 posts Page 1 of 1
In total there are 13 users online :: 2 registered, 0 hidden and 11 guests (based on users active over the past 5 minutes)
Users browsing this forum: Bing [Bot], Semrush [Bot] and 11 guests