Use of a Rheodyne 7725 valve with gas at high pressure
I wonder if some of you have experience with use of a inert gas mixture at P of 200 bar or above (and maximum limit) with a Rheodyne 7725 liquid valve or if you see any problems to use a liquid valve as a gas valve on these conditions?
Many thanks in advance
Adão Araújo
Portugal :)
Ask Rheodyne.
My only concern would be the lower viscosity would allow gas to creep through crevices, so ensure the valve is in good condition and all bolts have been tighened to ensure really good contact, and switching is performed rapidly.
If possible, try to allow the valve to sit at low pressures when not being used to reduce chances of gaseous tracking.
Bruce Hamilton
Thank you Bruce,
I wonder if you know other alternatives besides Rheodyne. I'm also asking them anyway and hope I can have an answer.
6 way valves in our market are not so common and our only source is Rheodyne.
Adão Araújo
You can get 6 port valves for GC use from Valco instruments. Their agent in Portugal is as follows.
Paralab LDA
Phone (+351) 22 466 43 20
Fax (+351) 22 466 43 21
Gasman
Thank you for your answer.
I need valves for HP till 200 bar and above. I know that VICI also has this valves but limited in P rabnge till 60 or 80 bar. Sure I can use their valves for this purpose?
Adão Araújo
You should contact Valco. They will also need to know the temperature that you will be working at. I have supplied gas sampling valves from Valco that go to 5000 psi (344 Bar).
Gasman
In total there are 19 users online :: 1 registered, 0 hidden and 18 guests (based on users active over the past 5 minutes)