Anyone familiar with Nucleosil S3W column use?
Hi All,
Has anyone used Nucleosil S3W before? Can I use polar solvents on it. What are the limitations?
Thanks!
I`m quite sure that there is no such column. Where did You hear of it or where did You get it from?
My apologies, I was looking through several different columns and made a mistake. You are right that there's no Nucleosil S3W. The right name of the column is "Spherisorb S3W"
Waters Spherisorb S3W is a silica column. It can be used in any solvent that you like, except in the alkaline pH in aqueous solvents.
As every silica column, a separation may be sensitive to the water content in your mobile phase, if you use a very nonpolar solvent or solvent mixture such as hexane. If you use polar additives in your mobile phase, the sensitivity to the water content in your mobile phase decreases significantly.
In total there are 23 users online :: 2 registered, 0 hidden and 21 guests (based on users active over the past 5 minutes)
Users browsing this forum: Ahrefs [Bot], Bing [Bot] and 21 guests
