Posted: Thu Nov 01, 2007 12:14 pm
This is driving me nuts! i have been running blanks for weeks and i can't get rid of a persistent benzene peak. i am running a Tekmar 3000 P&T concentrator with a 6890/5973 GCMS, and an EST autosampler with automatic standard injection.
I first sonicated the valves on the concentrator in methanol, changed the trap and flushed the transfer line with methanol. - still benzene.
I ran samples using a blank trap - still benzene.
I then ran samples through the glassware. - no benzene, so I turned my attention to the autosampler. I ran samples without standard addition - still benzene. I changed the water reservoir and filled the standard vials with pure methanol just to be sure. i also changed the pencil filter for the water supply. still benzene.
i have clipped the end of the transfer line from the autosampler to the concentrator and that does'nt seem to have made much difference.
Next thing i will try is to purge a vial of methanol to flush the system.
Does anyone know of any components that would release benzene into the system?
thanks
I first sonicated the valves on the concentrator in methanol, changed the trap and flushed the transfer line with methanol. - still benzene.
I ran samples using a blank trap - still benzene.
I then ran samples through the glassware. - no benzene, so I turned my attention to the autosampler. I ran samples without standard addition - still benzene. I changed the water reservoir and filled the standard vials with pure methanol just to be sure. i also changed the pencil filter for the water supply. still benzene.
i have clipped the end of the transfer line from the autosampler to the concentrator and that does'nt seem to have made much difference.
Next thing i will try is to purge a vial of methanol to flush the system.
Does anyone know of any components that would release benzene into the system?