what could be the problem??
Posted: Mon Nov 19, 2007 3:47 pm
by che313
Dear everybody
This chromatogram is from a new 5975B Agilent MSD
what you are think the resaons for this?
the column is HP-5ms (30m & 0.25mm)
Posted: Mon Nov 19, 2007 3:58 pm
by zokitano
Dear Che313,
You should give more information about the method if you expect good response!
Be more specific about oven temperature program, injector and detector parameters, carrier gas parameters, type of analytes involved, what are you trying to achieve etc.
Regards
Posted: Mon Nov 19, 2007 7:38 pm
by bhuvfe
Sample too concentrated? dirty inlet?
Posted: Tue Nov 20, 2007 12:11 pm
by che313
Dear Che313,
Regards
oven:
initial 200 C
then 25 C/min to 280 C hold 4.8 min
inlet 250 C
MS Quad: 150 C
MS Source: 230 C
carier: He 1.0 ml/min
target: d9-thc (tetrahydrocannabinoid)
dervetise: MSTFA
Posted: Tue Nov 20, 2007 3:05 pm
by che313
it could be
which chemical I have to run to insure my GC is functuned properlly?
Posted: Tue Nov 20, 2007 6:50 pm
by AICMM
Che313,
What column are you using, what solvent are you using? Are you running SIM or scanning? If scanning, can you post a spectrum, if SIM what ions? How does your tune look? Still need more data.
This could very easily just be junk from your sample but it is hard to say without more information. Your MS came with a test mix (typically octafluoronaphthalene since it has a great EC capture coefficient) so try running that under the recommended conditions and see what that looks like.
Best regards.
Posted: Thu Nov 22, 2007 12:42 pm
by che313
Che313,
Best regards.
thank you for your attention
the coulmn as mensioned is HP-5ms
solvent is ethyl acetate
scan mode
tuning is good - same as supplied with ms
we didn't got the test mix!!
is ethyl acetate good solvent for blank and for syring cleaning purposes?
do you suggest another kind of solvents for syring wash? ( now we use ethyl acetate pre and post injection 2 times)
Posted: Thu Nov 22, 2007 4:08 pm
by Consumer Products Guy
Looks like a GC issue, not an MS issue. Why not start with something really simple, like methyl laurate or dodecane dissolved in solvent? Anthing you have that's relatively pure, non-polar and maybe 150 to 350 molecular weight......
Posted: Thu Nov 22, 2007 4:33 pm
by zokitano
Che313,
You said that you're analysing THC derivate. Did you extract it from biological samples? If yes, are you sure that sample preparation method was suitable? Maybe you have interfering compounds from your sample.
You said also that you used derivation of your target compound. Are you sure that the derivation reagent is pure enough? Derivation is additional step where you may contaminate the samples.
Did you run blank to see whether your solvent is free of contaminations?
Regards
Posted: Fri Nov 23, 2007 12:08 pm
by che313
Che313,
Regards
yes its extracted from urine. although the prepration method has a refernce, it can't be said it will be clear 100% from interfering. And also I can't garntee the dervative is clean 100%.
If those reasones are affecting I will be satisfy, because I was thought the problem from GC or MS it self. for that I was asking about a reference material I have to run to check the functionality of my GC.
Posted: Fri Nov 23, 2007 3:03 pm
When you say you are derivatizing, do you mean making trimethylsilyl (TMS) derivatives? If so, do several of the peaks shown have significant 73 m/z ion? If not, you may not have enough derivatizing agent in there to react with remaining water and have excess to react with your THC. Maybe try more derivatizing agent.
