A quick question...how is it possible to have recovery consistently between 102-110% when spiking serum samples?
I am spiking with the highest standard. The standards are made in spiked bovine serum albumin PBS buffer...The standard curves are nice and linear.
Just curious...anyone got an idea?
Plasma can both enhance and suppress the response of an analyte. It appears that you caught a case of ion-enhancement...
I can understand that if the internal standard is slightly different. But is it really possible that ionenchasement/suppression can produce 108% recovery when doing isotope dilution ms/ms?
Why not? The enhanced or suppressed response can be caused by the mess between MS1 and MS2.
Is the blank clean - i.e. no peak at the traget retention time ?