Planning of a GC lab gas supply
Posted: Sat Feb 02, 2008 8:50 pm
After much (almost unholy) delays and strange issues I can finally start moving into my new lab.
While mounting the first tables I stumbled over a small but crucial issue: How do I bring the gases to my GCs? I have around a dozen of GCs (Varian 3700 and some 3400 as well as DANI boxes I rebuild almost from scratch).
I have a wide area of applications where the focus is mostly on trace analysis and reliability. My employer is not very forgiving.
I'll use Nitrogen, Hydrogen and Helium with the option of Argon/Methane for an ECD in the future.
As I can setup my lab only once (for practical and financial reasons) I have several questions to you:
1.) I'd use chopper gas lines (purified for GC purposes of course) as they are cheaper than stainless steel, easier to plump, immune to hydrogen induced corrosion (HIC) and the material is compatible with the inlets of my GCs (stainless steel is harder than chopper and could damage the seat of the receiving fitting in the GC).
So is there any compelling reason to use stainless steel gas pipes?
2.) Should the gas lines be hard soldered (my dad could do that for me) or are Swageok/Gyrolok/... fittings the better choice? My lab should stay (more or less) untouched for the next 20-30 years
3.) I plan to place the gas bottles about 20 meters away from my GCs so what internal diameter should my gas lines have and is it good practice to install a loop of gas piping to damp pressure pulses?
4.) Is it beneficial to install small loops of gas plumping directly at the GC as another means to damp pressure drops?
5.) Regarding the gas quality I chose 5.0 purity all over. When using Helium that purity gets expensive. According to Supelco one can purify 4.5 helium to around 6.5 with the 3800-U heated gas purifier. The data sheet looks interesting and the saving could be substantial. Do you have any experience feeding an Ion trap mass spectrometer with gas purified that way?
6.) I heard that many GC labs use purified air as make-up gas. The savings may be (somewhat) tempting (less plumping, less cylinders etc.) but I see safety problems here (hydrogen flame could under some circumstances backfire into the air line). So what experiences do you have?
7.) Is there a cheap and reliable way to prevent contamination when changing gas cylinders?
8.) What measures are sensible to protected the lab from havoc caused by the GCs? I already installed fire detectors and a hydrogen detector which sniffes gas on the top of the ceiling. In case hydrogen is detected the hydrogen line is to be switched over to nitrogen, in case of a fire alarm all GCs (except the MS) should just be switched off.
Nonetheless there are massive objections to let the GCs run unattended. Which additional measures are sensible to minimize the risk?
Thanks for reading. I hope you can help me a bit