How to resolve the noise in Chromatography spectrum in L51 ?
For the title:
I encountered some problem in experiment of USP Tadalafil for Enantiomer and Diastereomer purity.
In below control and condition:
Mode : LC
Detector : 222nm
Column : 4.6-mm X 25-cm; 10-micrometer packing L51
Flow rate : 0.75 mL/min
Injection volume : 10 micro-liter
In the spectrum of Sensitivity solution, I found a tiny peak at retention time of 4 min. Because of the concentration of this solution was too thin to pass the requirement of sensitivity. The signal-to-noise ration must be not more than 20.
Do you have any suggest for me to refer?
Thanks.
Regards
Hello.
Was the system flushed with isopropanol after reverses phase chromatography before normal phase one? Try to mix mobile phase in single bottle, degas it via ultrasound, disable other pump channels.
Best regards,
Dmitriy A. Perlow
In total there are 17 users online :: 1 registered, 0 hidden and 16 guests (based on users active over the past 5 minutes)
Users browsing this forum: Semrush [Bot] and 16 guests
