Gradient performance tests
I'm thinking of changing our bi-annual HPLC system performance checks by incorporating the gradient performance tests described by John Dolan in a July 2006 issue of LCGC. Can anybody who uses these tests tell me the acceptance criteria they set for each test (linearity and step-wise tests)?
Hi carnie,
You can divide the linear part of the test/data into 5 points by placing a drop-line across the linear range. Then use the calculated height as Y- values and the time (minutes) as X- values. The acceptance criteria could be: r = 0.99 and visual linearity confirmation.
The step-part could be handled as individual data points. Calculate the mean and the standard deviation for all the steps and the acceptance criterion could be: RSD -less or equal 5%.
In total there are 6 users online :: 3 registered, 0 hidden and 3 guests (based on users active over the past 5 minutes)
Users browsing this forum: Bing [Bot], Google [Bot], Semrush [Bot] and 3 guests