Calibration Not Calculaing Expected Concentration Properly
I prepared 5, 10, 50, 100 PPM standard from pure ethylene glycol ran it on GC-FID. When i went and did quantification of calibration. The concentrations are not right. Something weird is going on in the software. Can anybody help please. For 5ppm it is giving 17ug/L. For 10PPM it is giving 9.5 ug/L. For 50 PPM it is giving 17ug/L for 100 PPM it is giving 115ug/L. I tried everything i don't know what am i doing wrong. Can anybody help me in this situation. Thanks
Could be a lot of things. Did you populate your component editor correctly? Did you maybe use the dilution/sampleweight fields when you shouldn't have? Typically the concentrations are input into the component editor.
Also, did you set up a sample set with wizard, are you sure you set the standards to different levels per vial instead of the same level?
What have you tried?
Agreed
Check calibration standards and curve linear
Go to alter sample table and check component table
Create new processing method if nothing obvious
Thank you very much for your responses got it working now it seems my calculations were wrong
what was the issue here ?
It might be useful to share your miscalculation in empower
In total there are 25 users online :: 0 registered, 0 hidden and 25 guests (based on users active over the past 5 minutes)
Users browsing this forum: No registered users and 25 guests
