vials advice in GC-MS

Posted: Fri May 09, 2008 8:13 am
by pesquera
Dear all,

I work with a thermo TraceUltra-DSQII and last week we had a several contamination due to vials (septum of caps).
I started in April with a grant and the GC-MS is new.

Could you help me with the selection of new vials? (material and trade mark)

Thank you,

Posted: Fri May 09, 2008 12:11 pm
by JGK

Posted: Thu Jul 10, 2008 3:31 pm
by Ron
With the newer GCMS systems it is common to see light siloxanes from septa. This is more of a problem if mulitiple injections are made from the same vial. The problems tends to be worse wtih volatile solvents like methlyene chloride. One common source of the siloxanes is the rinse vials if they have septa. Change the rinse solvent often, and if the rinse vials have septa they should also be changed often. The best rinse vial tops I have used are the diffusion caps for the 4 mL vials that Agilent uses. I don't know if other companies sell these, but they reduce solvent evaporation without the use of a septum, and greatly reduce siloxanes in the rinse solvent.