Is it possible to have a lingering 'memory effect' in a HILIC column even after prolong flush - ie 95% water and then 95% ACN and then 95% water again for approcimately 10 hr?
Everytime I have run a batch of app. 80 samples, I started to experience this memory effect in my column for the peak of interest that came out at the app. RT even when I injected a blank sample.
How can I go about removing this problem? There is no issue if I change to a new column ( obviously, I looked like there is some memory effect from the column ).
My gradient: Water + 10mM amm. acetate, ACN + 10mM amm. acetate
Total run time: 10 min
Flow: 0.25mL/min
Column: 150mmX 2.1mm X 3um HILIC column
Gradient conditions: 97 ACN to 80 ACN in 3 min, then 5 ACN for 0.5min and back to 97 ACN for 6.5min. Equilibration time: 5 min
Do advise if you have any idea. Thanks a lot.