vials advice in GC-MS

Discussions about GC-MS, LC-MS, LC-FTIR, and other "coupled" analytical techniques.

3 posts Page 1 of 1
Dear all,

I work with a thermo TraceUltra-DSQII and last week we had a several contamination due to vials (septum of caps).
I started in April with a grant and the GC-MS is new.

Could you help me with the selection of new vials? (material and trade mark)

Thank you,

Good judgment comes from bad experience, and a lot of that comes from bad judgment.

With the newer GCMS systems it is common to see light siloxanes from septa. This is more of a problem if mulitiple injections are made from the same vial. The problems tends to be worse wtih volatile solvents like methlyene chloride. One common source of the siloxanes is the rinse vials if they have septa. Change the rinse solvent often, and if the rinse vials have septa they should also be changed often. The best rinse vial tops I have used are the diffusion caps for the 4 mL vials that Agilent uses. I don't know if other companies sell these, but they reduce solvent evaporation without the use of a septum, and greatly reduce siloxanes in the rinse solvent.
3 posts Page 1 of 1

Who is online

In total there are 26 users online :: 0 registered, 0 hidden and 26 guests (based on users active over the past 5 minutes)
Most users ever online was 11462 on Mon Dec 08, 2025 9:32 pm

Users browsing this forum: No registered users and 26 guests

Latest Blog Posts from Separation Science

Separation Science offers free learning from the experts covering methods, applications, webinars, eSeminars, videos, tutorials for users of liquid chromatography, gas chromatography, mass spectrometry, sample preparation and related analytical techniques.

Subscribe to our eNewsletter with daily, weekly or monthly updates: Food & Beverage, Environmental, (Bio)Pharmaceutical, Bioclinical, Liquid Chromatography, Gas Chromatography and Mass Spectrometry.

Liquid Chromatography

Gas Chromatography

Mass Spectrometry