Method development (wavelength choice/RRF of impurities)
When selecting a wavelength for early drug development (stage 1), it can be assumed that the impurities have a RRF of 1. I remember reading this in some FDA guidelines.
My question is is there any criteria that must be met in order to make this assumption? ie can this assumption be made if detecting at a wavelength that is not a maximum?
In other words if my compound has a maxima at 238nm, but the detection wavelength of the method is 252nm would the RRFs of the impurities scale linearly? thx
Hello. Check for Absorbence 1 to Absorbence 2 ratio for different peaks to answer your question.
Best regards,
Dmitriy A. Perlow
That's a good idea. I don't know why I didn't think of this. thx
The original assumption might just as well say "it can be assumed that the impurities have the same chromophore as the product", which is obviously a terribly bad assumption.
If the absorption spectrum for the impurity is the same as that of the product, then if you measure the product at a slightly suboptimal wavelength, the measurement of impurity will be equally suboptimal, and the calculation of percentage purity will be unaffected.
If the absorption spectrum of the impurity is different to that of the product then it's pretty much certain that its molar extinction coefficient will be different too (except, by purest accident, at some particular and unrecognisable wavelength), and in this case your calculations are, in any case, meaningless in an absolute sense; they can only used in a relative sense, to establish whether this batch is better than that batch. If working in a relative sense, again, it does not matter what wavelength you use for each.
There's no point in worrying too much about data based on dodgy assumptions; worry can't make the data better. It's more productive just to understand, and accept, the data's limitations. Not that the authorities will necessarily agree.
