the unit of slope in calibration curve
calibration curve is constructed by plotting area of each peak vs corresponding concentration. I'd like to know what would be the unit of slope. Is mAbs*s/ugml-1 correct?
More generally, it is the ratio of the y-axis units to the x-axis units.
Peak area (y-axis) is most generally voltage X time, and so could be
millivolt-sec or
microvolt-sec (or any other combination of voltage and time units).
If the data system has done an intermediate calculation to absorbance, then the units would be
AU-sec or
mAU-sec.
Concentration (x-axis) could be expressed as mass/volume and so could be
mg/mL, or
g/L, (or any other combination of mass and volume units).
It could also be expressed in molar terms, in which case the units would be moles/volume.
Just to complicate life, the x-axis could also be expressed as mass-on-column, in which case the units would simply be mass. :?
So, the response factor (slope) could be mAU-sec-mL/μg (which is equivalent to what you wrote), or it could be:
μV-sec/μg ,or
mAU-sec-mL/mol , or . . .
It all depends on how your data system and inputs are set up.
Users browsing this forum: Bing [Bot] and 14 guests
