NP sub 2 um columns
Are there suppliers of NPLC columns on bare silica? Particle size can not be lager than 2.5 um.
What is the big difference between a cyano/phenyl NP mechanism and a bare silica column?
Why can the particle size be no bigger than 2.5um?
Most suppliers will do columns down to 3µm
In normal phase chromatography cyano/phenyl columns are generally less affected by the presence of water in the mobile phase. They will also re-equilbrate faster between injections for gradient methods.
Hi Koen,
If you need a sub 2μm particle column, why don’t you try the Acquity HILC column?
It might be able to do the job for you.
Because we want to speed up the analysis. The “fastâ€
Knauer in Berlin seems to produce NP-columns using 1.8 um material. I remember that I saw somewhere on their website a Tocopherol NP-application in less than 2 minutes.
In total there are 14 users online :: 0 registered, 0 hidden and 14 guests (based on users active over the past 5 minutes)