I have ran into difficulties with one of my methods, and need some help to figure out what's going on. Simply speaking, when spiking biological samples and calculating the recovery, I find 120-130 % - not 100% as I expected.
Everything else looks fine. Excellent precision and linearity of standards. Linearity of spiked biological samples quite perfect. Separation from known interfering substances is good and adequate. Peak shape is symmetrical, a little narrow perhaps (12s), but the MS acquire at least 16-18 measurements during this time.
The standard curve samples are prepared in a pseudomatrix (charcoal-stripped) - although great care has been taken to remove all charcoal.
The internal standard is a deuterated isotope of the steroid hormone I aim to quantitate.
Samples are prepared by simple dilution in H2O, before it's injected onto an Oasis HLB trap column. After a washing procedure with MeOH:H2O, it's eluted by an ACN:H2O gradient to an C18 column.
Now...the obvious source of errors is the standard curve, but preliminary tests indicate this is ok.
So...my thought are now on a the MS. I monitor two fragments in MRM, both show the same "overestimated recovery". The IS share one daughter ion with the analyte, but mother ion are 4D heavier - and no cross talk occurs.
Is it still possible that some kind of ion enhancement may occur between Q1 and Q3? Could there be something in biological samples that either:
1. increase the ionization of the mother ion of my analyte compared to IS?
2. increase both fragments of my analyte compared to the fragment of IS?
Is this even theoretically possible? Have anyone experienced such phenomena? Should I just forget about this "wild" though, and start over again troubleshooting my standard samples?
Thankful for advice!
Anna
PhD student