Reproducibility Problem
I have problems of reproducibility, in an Alliance 2695, have mounted 3 systems of 6 injections, and always in the three systems the first injection is the one that does not play like other 5. im not have idea what it is.
It's hard to give a specific answer without a lot more information about your method, but just as a guess: If your separation involves a gradient, the pre-equilibration time can have an impact on results. In an automated system, the equilibration time between injections is constant *except* for the first injection of the day. My practice has been to always make the first injection a blank or "dummy" to avoid that problem.
We ALWAYS did that; likely learned that from you in the late 1970s !!!
We wrote that into our test procedures so there would be no confusion, and for our cGMP compliance.
Tom Jupille and Consumer Products Guy:
Great Advice! I must have learned this from reading your advice over the years.
I generally include 1 - 4 equilibration injections of standard prior to system suitability to condition the system, depending upon the assay.
Also, I suggest adding time to the re-equilibration at the end of the chromatogram or between injections, if possible.
Users browsing this forum: Bing [Bot], dluk and 4 guests