The result of my autotune is mysterious.
The mass 69 give 420 000
The mass 219 give 148 000 (35% of 69)
The mass 502 give 15 000 (3.5% of 69)
Habitually, i have
69: 100%
219: 80 - 100%
502: 8 - 10%
My reppeler is high, 34.8 but my source is clean.
What do you think about this problem?
Thank you
How about the isotopes?
Has calibration compound been used for very long?
Hi!
Isotope ratio is o.k......
70: 1%
220:4.6%
503: 10%
Hi Rick!
Can you post the Autotune Report and the Tune Evaluation Report?