Chromatography of Zwitterionic Buffers (MES, MOPS, etc.)
5 posts Page 1 of 1
Hello:
I am looking for a simple method to analyze zwitterionic buffers, like MOPS and HEPES
http://en.wikipedia.org/wiki/MOPS
What are my choices? I have no retention on RP column even at very low organic (various pH)
That would be a classic application for ion-pair chromatography, or else one of the "mixed-mode) packings with embedded ion-exchange groups. Or, you could use good old ion-exchange (just run at a pH that suppresses one of your charged sites).
I would consider HILIC as well.
I found this article:
Analytical Chimica Acta 583 (2007) pp 124-127 where HEPES is the internal standard and they used an amide-80 column.
If you found any other solutions, I'd be interested in hearing about them.
Here are few methods for several zwitter-ionic buffers:
http://www.sielc.com/compound_256.html
regards,
Vlad
Vlad Orlovsky
HELIX Chromatography
My opinions might be bias, but I have about 1000 examples to support them. Check our website for new science and applications
www.helixchrom.com