column with benzene contamination
Posted: Fri Sep 27, 2019 7:37 pm
by lablab15
Our column is contaminated with benzene, I would like to know what we can do to remove it.
We have already tested possible causes such as: syringe, injector, solvent ...
Thank you
Re: column with benzene contamination
Posted: Fri Sep 27, 2019 9:49 pm
by James_Ball
What column are you using? Packed or capillary? Phase? Size? Carrier gas?
Different columns may need different solutions to the problem.
Posted: Fri Sep 27, 2019 10:03 pm
by Consumer Products Guy
Won't that just volatilize and get swept out by the carrier gas at temperature?
Posted: Mon Sep 30, 2019 6:27 pm
by lablab15
The column is the ZB-624, L = 30m ID = 0.53mm.
Nitrogen gas such as drag, hydrogen and synthetic air for combustion.
solvents we use isoctane, hexane, dichloromethane, methanol ... but for this analysis we use methanol.
Posted: Mon Sep 30, 2019 6:51 pm
by James_Ball
If baking at maximum temperature for 20 -30 minutes does not remove benzene from that column, it must have some serious non-volatile contamination present.
Have you remove the first 1 or 2 loops of the column? There may be some contamination that adsorbs benzene very well and only releases it when an injection is made and the solvent removes a little each time. Otherwise this phase should not retain benzene if baked well.