Column Bleed?
I'm not experienced with HPLC and was flushing my system (Agilent 1100) to get a baseline when I saw this.
Red: Pressure, Blue: VWD 254nm
Column: BDS C18
Mobile phase: 50:50 Methanol:Water
Flow rate: 0.8 mL/min
Is that due to column bleed?
Since the four negative dips in pressure are closely followed by four baseline excursions, it looks more likely that you had momentary losses of prime (outgassing from the solvents?). Difficult to tell more without being able to read the y-axis units. Any "bleed" from the column would be negligible, and very un-likely to show up at 254 nm.
I degassed the solvent for 15 minutes prior to use, so I'm not too sure what else can cause that.
Today, I experience the same thing:
Column: BDS C18
Flow rate: 0.5 mL/min
What else can cause such observations?
The only thing that can cause what you are seeing is a temporary "loss of prime". Most likely possibility is an air bubble, as discussed (more on that below). Another possibility is a sticking or malfunctioning check valve.
Re degassing, some questions:
1. Are you pre-mixing your mobile phase, or are you mixing on-line?
2. If you are mixing on-line, which version of the 1100 do you have? the low-pressure mixer (a single high-pressure pump with a proportioning valve), or the high-pressure mixer (two high-pressure pumps)?
3. How, exactly, did you degas?
In total there are 25 users online :: 2 registered, 0 hidden and 23 guests (based on users active over the past 5 minutes)
Users browsing this forum: Ahrefs [Bot], Bing [Bot] and 23 guests
