Problem with the MS 5973: Difficulty with the EM supply
Hi people!
I work at a GC/MSD (6890/5973).
This system don’t has an UPS, and after some interruptions of electrical supply that began to fail.
This is the error message in the Logbook:
1101, "[{REP} acq: strt sim]: Unrecoverable mass spec fault: 512"
Difficulty with the electromultiplier supply at runtime 7.44
This problem had happened beforehand, but the failure was in the method, since power failures ocurred during a run, and the method was corrupted.
Today, I executed a run, and the next run stopped at 7.44 minutes.
The filament is good and the vacuum is good too. The EMV was 1500.
I executed Profile and Scan (in Manual Tune) and the error message is the same.
Help me please!
Astroguille,
Probably is worth to use a different method to see if you get
the same error message and also to retune the instrument using a
previously saved copy of the autotune macro (or a macro from another
instrument or computer).
As you mentioned something similar already happened in the past...
Best regards,
bhuvfe