creatine monohydrate
Dear all,
I am currently trying to determine the LOD for creatine monohydrate. Mobile phase composition: KH2PO4 monobasic adjusted to pH4 using phosphoric acid, isocratic condition. Wavelength: 210 nm. Column: Waterssymmetry shield 4.6 x 250 mm , 5 um.
So far, I am able to get the peak at around RT of 2.9 min as well as 3D spectrum at around 195 nm at higher concentration from 25 ug/ml to 5 ug/ml. But when I tried running at lower concentrations say below 1 ug/ml, I noticed that there isn't any change in peak heights at the different lower concentrations (0.5 ug/ml, 0.1 ug/ml, 0.05 ug/ml etc. ). At the same time, I am also unable to get its respective 3D spectrum. Can anyone know why this is so?
I wanted to find out if the analyte is a basic or acid or neutral compound but not sure how to go about doing it. Is there any particular website that I can refer to?
Also, if I want to know their UV cutoff for both my analyte and the buffer that I am using, where can I locate such informations?
Appreciate if anyone can give me some advise. Thanks. :)
Tlili,
You might have another compound eluted at the same time or you are maybe too close to the void volume (if not exactly at it).
Have you tried to make a blank injection to see how your chromatogram looks like?
There is a prety good pKa compilation for your and other compounds that can be found at: http://research.chem.psu.edu/brpgroup/p ... lation.pdf
Your mobile phase is quite transparent to UV. Creatine should absorb in UV due to it's carboxylic group so it should adsorb in low UV wavelengths (although I do not have it's structure in mind).
Also: Be careful, phosphate is not a buffer at that pH, and creatine is not very stable, for instance, it may convert to creatinine. Kostas already pointed out that you have a good indication that your peak is not creatine.
For UV info check catalogs, google, Merck Index, Stadtler ......
Just noticed that Stadtler should be Sadtler (Bio-Rad) instead, usually found in university chemistry libraries.
Also, sometimes one finds something there:
http://webbook.nist.gov/chemistry/
http://www.colby.edu/chemistry/cmp/cmp.html
In total there are 24 users online :: 0 registered, 0 hidden and 24 guests (based on users active over the past 5 minutes)