Problem with Diphenylamine
2 posts Page 1 of 1
Hi people!
I analize pesticides (Multi-Pesticide Residue Analysis) in fruits, vegetables, water and soil with an Agilent GC6890/MSD5973. I use a column HP5ms.
I have a problem with Diphenylamine: this pesticide appear in almost every samples (in blank samples too).
I changed the septum and the liner, without success results.
The system is contaminated, but I don't know where.
Please help!
astroguille,
Because of the blanks, start with solvent. Go to a different one, not the same type from a different vendor but a different solvent.
Then, you name it, syringe, waste vials, rinse vials, rinse solvent, split vent line, front of column,.....
Best regards.