Hi people!
Does anybody have a method for determination of methyl bromide in oranges?
I suppose that can do it by Headspace.
Thanks!
Given that it's a regulated pesticide, contact your national food safety authority for details of the approved methods.
A Google search using "analysis of methyl bromide in food" revealed plenty of methods, headspace and otherwise.
Bruce Hamilton
astroguille,
Headspace should work very well. ECD would be my detector of choice. A long volatiles column, or a thick film -1 type, or a Plot Q would be my starting point for columns.
Best regards.
This is also a routine analyte by EPA method 8260 using Purge and Trap which can handle soil/solid samples. The method is located at:
http://www.epa.gov/epawaste/hazard/test ... /8260b.pdf
Greg
Thank you people!