Passivation
8 posts Page 1 of 1
I was reading a book about HPLC and the passivation with HNO3 6M was recommended once a month.
It is the first time I was hearing about this. I am afraid to use it.
Does anyone use this technique ?
Thank you
are you sure it was written "once a month" ?
I didn't use it and never got any problem
Yes I am sure. I still have the book.
We have agilent and shimadzu equipments and no where in the manuel this technique is mentionned
Once a month is really too much. Unless you have an application that creates
a very noisy baseline on your detector afters such a short time.
We use it only when needed, i.e. to clean a refractive index cell or a diode-array cell.
I would use it only if my application requires it or for cleaning purposes...
thank you
and do be careful, anyone reading this thread. In a building less than 100m from where I write, someone once blew up a winchester bottle of hplc waste by pumping nitric acid during passivation into the bottle which already contained isopropanol.
Before anyone denies it's possible, it really happened, and there's a UK health and safety executive report on this incident somewhere.
I did it and the system recovered a fairly good sensitivity, if I use a stronger mobile phase. I still can see some masses on the scan of the mobile phase, that were not there before.
The system was very dirty so I had to disconnect and use the ultrason for some fittings. They came out as new.
The system was so dirty that I am not sure that the mixer and the pulse dampener are clean again. I don't know how is built a mixer and a dampener. Could they be put in the ultrason bath with acid, would it do something good.
They have used tropolone in the mobil phase to complex organo-metallic
And it is every where
8 posts Page 1 of 1
Users browsing this forum: Bing [Bot] and 22 guests
