Reputable GC-MS repair services near Chicago?
3 posts Page 1 of 1
Hi everyone,
We have a Agilent 6890/5973N GC-MS that needs servicing and MS source cleaning. Are thee any service techs or companies that can do this in the Chicago area? We would prefer to have someone come onsite or we can take the system to them in person.
Apologies ahead of time if I posted this in the wrong sub-forum.
If it just needs a source cleaning that's something you can do on your own. I've seen people reassemble the source with the spacer in front of the drawout plate and the instrument still ran. As long as you don't jam it back in full of alumina you should be fine.
https://community.agilent.com/technical ... /files/847
Try contacting Geoff Wilson at Great Lakes Technical. They are in Milford, MI.
https://www.greatlakestechnical.com
He is who we had do our upgrade of a 5973 from HPIB to LAN a few years ago.
3 posts Page 1 of 1
In total there are 7 users online :: 1 registered, 0 hidden and 6 guests (based on users active over the past 5 minutes)