LCSolution 1.25: Identification but no quantification
Discussions about chromatography data systems, LIMS, controllers, computer issues and related topics.
2 posts Page 1 of 1
Hello all,
I'm currently facing issues with my HPLC where it is identifying the peaks correctly but is not quantifying them to give results (0 g/L for all peaks). I would really appreciate it if someone could help me understand the underlying problem.
Please find the specifications of my post-run method below:
1. Integration
- Width: 5 seconds
- Slope: 15000 uV/min
- Drift: 0 uV/min
- T.DBL: 0 min
- Min. Height/Area: 0 counts
- Calculated by: Area
2. Identification
- Window: 5%
- Peak Selection: Largest Peak
- Retention Time Update: Average
3. Quantitative
- Quantitative method: Internal standards
- Calculated by: Height
- Calibration Curve:
- # of calibration levels: 10
- Curve fit type: Quadratic
- Zero: Forced through
- Weighting method: 1/A
Only the guess:
None of the peaks marked as internal standard.
2 posts Page 1 of 1
In total there are 8 users online :: 0 registered, 0 hidden and 8 guests (based on users active over the past 5 minutes)