high initial temperature
Hi All,
we are using a CP-Volamine column for a separation of semi-volatile analytes (BP ~ 400°C). Their elution temperature is ~280°C in gradient elution. The isothermal limit of the column is 265°C and the gradient limit 300°C.
Right now we start the gradient at 150°C and ramp to 280°C.
Do you see a problem with having the inital temperature yet higher, e.g. 225°C for the gradient elution?
The ramp rate could then be somewhat reduced to effect elution of the components ON the gradient slope rather then directly behind it.
I only ever see 40/50°C initial temps or up to 150°C for semi-volatiles applications.
Thank you!
Try it !!! Then evaluate whether better/worse/same.
Back when I was working, "gradient" was only associated with HPLC; GC used the term "temperature programming"; maybe things are different now....
Hi Consumer Products Guy,
Thanks for the input! I was wondering about having the column permanently exposed to an elevated temperature.
You could just have a standby method at a lower temp that's constantly loaded with a run method that starts at a higher temp if you're worried about decreased column life. When the sequence is over it'll drop back to whatever method it was set to before the sequence started.
Users browsing this forum: Bing [Bot], Google [Bot] and 11 guests