NH4OH with PLRP, proteins/peptides
I'm interested in trying different mobile phase as a way to improve selectivity in a multi-step purification scheme.
Peptide standards look fine when run on PLRP with 0.1% TFA and a linear gradient 5-60% MeCN. However, peak shape dramatically deteriorated when 15 mM NH4OH was used in place of TFA. What would cause this?
Perhaps the equilibration time must be longer?
Loading sample on a column equilibrated with 5% MeCN 15 mM NH4OH
My sample volume is 0.4 mL in 10 mM NH4OAc + 250 mM NaClO4, 5% MeCN.
After loading the 0.4 mL sample on to this 1 mm ID PLRP column at 50 ul/min, how long will it take to re-equilibrate with ammonia before beginning the gradient?
In regard to pH changes and the effect on RP separations:
J Sep Sci. 2007 May;30(7):949-63.
Selectivity in reversed-phase separations: general influence of solvent type and mobile phase pH
In regard to what you're asking:
If your standards looked fine with 0.1% TFA, why did you switch to NH4OH? What pH is your mobile phase now? What column are you running?
One more thing - do you mean 0.4 microliters or 0.4 milliliters injected?