Calibration curve linearity validation
A question on confirmation of calibration curve verification.
In the past i have used r2 value greater than 0.999 to confirm linearity. I have received the follow statement from a cilent.
"The acceptance of the model must be verified by evaluating the relative bias between the back-calculated and nominal concentrations of the calibration standards. The use of the correlation coefficient is not recommended for confirmation of the regression model."
I am not sure how to apply this. i assume the following:
say i used nominal concenration of 2, 4, 6, 8 and 10 µg/ml stds and plot these to obtain the responses. Then re-plot with actual concentrations based on weight taken, i.e. 2.2, 4.1, 6.3, 8.1 and 10.0 µg/ml. Then compare both responses, expressed as a %.
You are correct in your interpretation of the statement, it is exactly where they ask for.
r2 isn't a good tool for confirmation of linearity. Do some search on the internet (lack of fit test for example) and you will found out why.
Ace
Try reading the chapter on instrumental regression in this:
Statistics and Chemometrics for Analytical Chemistry J. N. Miller, J. C. Miller
WK
I'm Sorry I Haven't A Clue - Just A Minute - The Unbelievable Truth
Users browsing this forum: Bing [Bot] and 17 guests
