the back pressure grow up so fast
Dear all
I am assay the impurity of a drug product,and use the
column:Agilent ZORBAX SB-C18,4.6*150mm,5um,
buffer:0.2%dibasic potassium phosphate adjust to pH6.0 with diluent phosphoric acid;
mobilephase A:buffer-ACN(95:5)
mobilephase B:ACN
and mobilephase follow as below:
min A B
0-3 100 0
19 79 21
20 100 0
25 100 0
this method come from PF.
when I have injected about 200 ,the back pressure grows up about 30bars,and the perfoemence lose,I clean the column with ACN,methonal and back-flush it,the back pressure still high,
there is hypromellose in my test solution,is it cause the high back pressure?I am so puzzled.who can help me?
if you have suggestions,you can give me Email,thank you very much!
it is the hypromellosee that's causing trouble. Get rid of it prior to loading the sample on the column.
A simple filtration might do the trick. The hypromellose is not in actual solution - it's a suspention.
thank you!
but i have filtrated the sample before injection.
OK. Filtered using wich kind of filters?
Btw. what is the sample solvent?
the solvent is buffer-ACN(95:5)
and the filter is oganic filter,0.45um
Try 0.22 um pore size filter. Also make sure the filter material is somewhat hyrophobic.
If that doesn't work you might like to consider SPE.
Best Regrds
In total there are 23 users online :: 3 registered, 0 hidden and 20 guests (based on users active over the past 5 minutes)
Users browsing this forum: Bing [Bot], Google [Bot], Semrush [Bot] and 20 guests
